- Which screw(s) to remove?
- M3Z_3.jpg (137.7 Kio) Vu 6575 fois
Problème alignement Wild M3Z
Règles du forum
Présentations, questions et débats sur les loupes stéréo et macroscopes
Présentations, questions et débats sur les loupes stéréo et macroscopes
- PierreH
- membre

- Messages : 3170
- Enregistré le : 13 sept. 2007 14:21
- Prénom : Pierre
- Localisation : Buzet-sur-Tarn près Toulouse
- Contact :
Re: Problème alignement Wild M3Z
Hello Tom,
it is absolutely forbidden to remove these screws : they keep aligned lenses and vertical slider of the optical zoom.
It would be a nightmare to adjust the whole zoom system !
I don't think that the screws that held your system are inside microscope housing :
because the two parts of the microscope housing (top and bottom) are kept together with 3 long screws that you access at the bottom side.
I think (in fact, I am pretty sure) that the black plate that hold the motor is screwed with the same 2 screws that hold the microscope housing to the microscope holder.
These 2 screws are circled in blue on the photo.
You have to remove the black rectangular part to see these 2 screws...
it is absolutely forbidden to remove these screws : they keep aligned lenses and vertical slider of the optical zoom.
It would be a nightmare to adjust the whole zoom system !
I don't think that the screws that held your system are inside microscope housing :
because the two parts of the microscope housing (top and bottom) are kept together with 3 long screws that you access at the bottom side.
But this is not what you need to do :
I think (in fact, I am pretty sure) that the black plate that hold the motor is screwed with the same 2 screws that hold the microscope housing to the microscope holder.
These 2 screws are circled in blue on the photo.
You have to remove the black rectangular part to see these 2 screws...
Microscopes Zeiss WL, CP, DIC, épifluo, épiscopie HD, obj. Neofluar Phase, Plan-Neofluar 63, Optovar, écl. LED Seoul P4 3W
Stéréomic. Leica MZ12.5 Combi 3 Planapo 1x 2x, OPD Planapo 10x, écl. Schott KL1500LCD. Leica M3Z Plan Type S
Terrain Open University McArthur LED, Nikon Naturescope Mini, Emoscop SME LED, Belomo x10
Photomacroscope agrand. Kaiser modifié, Luminar 16mm, Apo-Rodagon N 50mm 2.8, Nikon CF 10 & 20x Plan EPI
APN Canon 450D téléc. USB, Fuji X10 Raynox DCR-250
Stéréomic. Leica MZ12.5 Combi 3 Planapo 1x 2x, OPD Planapo 10x, écl. Schott KL1500LCD. Leica M3Z Plan Type S
Terrain Open University McArthur LED, Nikon Naturescope Mini, Emoscop SME LED, Belomo x10
Photomacroscope agrand. Kaiser modifié, Luminar 16mm, Apo-Rodagon N 50mm 2.8, Nikon CF 10 & 20x Plan EPI
APN Canon 450D téléc. USB, Fuji X10 Raynox DCR-250
Re: Problème alignement Wild M3Z
Thank you for your reply!PierreH a écrit :You have to remove the black rectangular part to see these 2 screws...
I've tried before to remove the black rectangular part without success. I'll try again to find any screw behind.
I think the big white gear-wheel is bolted to the black triangular plate, but then sealed by grey epoxy.
I've not tried to remove the epoxy, not to damage anything...
Re: Problème alignement Wild M3Z
I've marked the epoxied areas in red.
I've also marked the areas I suggest the long screws you mentioned, in blue.
I'm not quite clear how to remove the epoxy.
Maybe I use a hot soldering iron...
PS: The reason, why I want to remove the unused parts:
I want to have access to the objective to change it or use an additional clamp-on-type 1.6x converter lens
I've also marked the areas I suggest the long screws you mentioned, in blue.
I'm not quite clear how to remove the epoxy.
Maybe I use a hot soldering iron...
PS: The reason, why I want to remove the unused parts:
I want to have access to the objective to change it or use an additional clamp-on-type 1.6x converter lens
Modifié en dernier par wild_tom le 19 nov. 2017 13:43, modifié 2 fois.
- PierreH
- membre

- Messages : 3170
- Enregistré le : 13 sept. 2007 14:21
- Prénom : Pierre
- Localisation : Buzet-sur-Tarn près Toulouse
- Contact :
Re: Problème alignement Wild M3Z
It will be hard to remove this black part : good luck !
You are right about the "blue" screws placement
You are right about the "blue" screws placement
Microscopes Zeiss WL, CP, DIC, épifluo, épiscopie HD, obj. Neofluar Phase, Plan-Neofluar 63, Optovar, écl. LED Seoul P4 3W
Stéréomic. Leica MZ12.5 Combi 3 Planapo 1x 2x, OPD Planapo 10x, écl. Schott KL1500LCD. Leica M3Z Plan Type S
Terrain Open University McArthur LED, Nikon Naturescope Mini, Emoscop SME LED, Belomo x10
Photomacroscope agrand. Kaiser modifié, Luminar 16mm, Apo-Rodagon N 50mm 2.8, Nikon CF 10 & 20x Plan EPI
APN Canon 450D téléc. USB, Fuji X10 Raynox DCR-250
Stéréomic. Leica MZ12.5 Combi 3 Planapo 1x 2x, OPD Planapo 10x, écl. Schott KL1500LCD. Leica M3Z Plan Type S
Terrain Open University McArthur LED, Nikon Naturescope Mini, Emoscop SME LED, Belomo x10
Photomacroscope agrand. Kaiser modifié, Luminar 16mm, Apo-Rodagon N 50mm 2.8, Nikon CF 10 & 20x Plan EPI
APN Canon 450D téléc. USB, Fuji X10 Raynox DCR-250
Re: Problème alignement Wild M3Z
I've removed one of the grey epoxy seals by the soldering iron method. But there is no screw, my assumption was wrong.PierreH a écrit :It will be hard to remove this black part
I see no accessible screw nor possibly hidden screws I could accesss from the outside of the microscope...
Thank you.PierreH a écrit :good luck!
Re: Problème alignement Wild M3Z
This morning I found 3 small holes at the white gear wheel.
When I slided in a thin 1,5mm drill and turned the gear wheel, I found 3 positions (each 120°) the drill could be slided in a few extra millimeters - and the gear wheel was some kind of "locked".
So, I thought, there's a sort of threaded ring (with 3 holes) inside that holds the gear wheel to the black plate, that could be unscrewed by the locked gear wheel .
When I slided in a thin 1,5mm drill and turned the gear wheel, I found 3 positions (each 120°) the drill could be slided in a few extra millimeters - and the gear wheel was some kind of "locked".
So, I thought, there's a sort of threaded ring (with 3 holes) inside that holds the gear wheel to the black plate, that could be unscrewed by the locked gear wheel .
Modifié en dernier par wild_tom le 19 nov. 2017 13:20, modifié 2 fois.
Re: Problème alignement Wild M3Z
I slided 3 drills into the holes, found the locking position, slided them in to the stop, locked internal ring and turned the gear wheel counter clockwise.
There was a slight friction that decreased a bit while turning.
I turned about 10 turns counter clockwise, but friction didn't decrease more than it decreased after the first 1/4 turn.
The axial play of the gear wheel didn't increase (as it should, if there was a loose thread).
So something inside holds the ring from inside the microscope housing.
There must be a way to open the microscope housing...
There was a slight friction that decreased a bit while turning.
I turned about 10 turns counter clockwise, but friction didn't decrease more than it decreased after the first 1/4 turn.
The axial play of the gear wheel didn't increase (as it should, if there was a loose thread).
So something inside holds the ring from inside the microscope housing.
There must be a way to open the microscope housing...
Modifié en dernier par wild_tom le 20 nov. 2017 23:38, modifié 2 fois.
- PierreH
- membre

- Messages : 3170
- Enregistré le : 13 sept. 2007 14:21
- Prénom : Pierre
- Localisation : Buzet-sur-Tarn près Toulouse
- Contact :
Re: Problème alignement Wild M3Z
Hi,
I sent an email to a friend who knows very well Wild/Leica stereomicroscope...
I sent an email to a friend who knows very well Wild/Leica stereomicroscope...
Microscopes Zeiss WL, CP, DIC, épifluo, épiscopie HD, obj. Neofluar Phase, Plan-Neofluar 63, Optovar, écl. LED Seoul P4 3W
Stéréomic. Leica MZ12.5 Combi 3 Planapo 1x 2x, OPD Planapo 10x, écl. Schott KL1500LCD. Leica M3Z Plan Type S
Terrain Open University McArthur LED, Nikon Naturescope Mini, Emoscop SME LED, Belomo x10
Photomacroscope agrand. Kaiser modifié, Luminar 16mm, Apo-Rodagon N 50mm 2.8, Nikon CF 10 & 20x Plan EPI
APN Canon 450D téléc. USB, Fuji X10 Raynox DCR-250
Stéréomic. Leica MZ12.5 Combi 3 Planapo 1x 2x, OPD Planapo 10x, écl. Schott KL1500LCD. Leica M3Z Plan Type S
Terrain Open University McArthur LED, Nikon Naturescope Mini, Emoscop SME LED, Belomo x10
Photomacroscope agrand. Kaiser modifié, Luminar 16mm, Apo-Rodagon N 50mm 2.8, Nikon CF 10 & 20x Plan EPI
APN Canon 450D téléc. USB, Fuji X10 Raynox DCR-250
Re: Problème alignement Wild M3Z
Thank you, Pierre!
Merci beaucoup!
Merci beaucoup!